For new lots of QC, we just run QC 10x each level and make sure results fall within manufacturer ranges (and SD/CV are acceptable of course). For strips, we do QC 2x each level and a minimum of 3 patient checks on the new lot vs current lot and main lab analyzer.
Erin
The process we use is pretty much the same as what Erin's facility does.
Good luck!
Jacqueline
For the patient checks, we use 6 mg/dL and/or 10% from CLIA total allowable error (westgard has a bunch of tables under their CLIA & Quality tab). The meter to main lab analyzer is not required for waived Accu-Chek, but we only get two lots per year so we just do them anyway. You would not be out of compliance if you did not do them for new lots.