My hospital has had the Accuchek Inform II for a long time now. I came into the POCC position a couple of years ago. I'm reviewing our current standard for this POC testing. On our standard, it states "normal" ranges for age groups, but the manufacture has not established any ranges so I'm wondering where these ranges came from. I understand that it is a CLIA-waived test, but how does one go about establishing a "normal" reference range for the meter.

For hospitals who use Accucheks, what "normal" ranges do you go by? Have you validated ranges or do you use your lab's "normal" ranges?

Critical values are already established per the hospital's lab.
Another question is how does one determine the linearity. I understand that the manufacture states the meter can read from 10-600. Our current linearity states that it is 40-500. This was before my time, and I'm still new to this. Was this established due to extra testing so that we didn't go with the manufacture linearity reading range?

I'm confused with a lot of questions.

5 Replies

We use our laboratory's normal ranges and verify for POCT.

When your hospital began using the AccuChek someone should have done comparisons between methods (Lab vs AccuChek) to confirm that the results were comparable.  This is generally done with at least 40 samples spanning the entire reportable range.  Perhaps your predecessor was only able to verify that results were linear from 40-500.  You should search for the results from the initial AccuChek validation studies.

Twice a year you should run a small number of comparisons (5-10) to verify that the results are still comparable.  If the laboratory method/instrument is ever replaced, then a complete validation must be done again.

I'm glad you were brave enough to ask these questions.  The community on this site is amazing.  Good luck, and welcome to POCT!

Thank you for the reply! 40 samples as in actual patient samples or would the accuchek linearity kit suffice? Currently, I'm performing the accuchek linearity kit on all my accucheks twice a year. I do not believe any of the accucheks have been correlated to our lab's instruments since the initial validation. Is that necessary?

We use normal ranges established by the American Diabetes Assoc. in all areas.

Hi Mai! For the linearity, we use the linearity kits that Roche provides, but don't hit the lower range to take you to the full 10-600. So, we also use patient samples that span the range and correlate with our main lab analyzer. Happy to help if you have questions erin.norberg@bmc.org

Thank you everyone who replied. You were all a great help!

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Mai Vang
about 4 years ago
5
Replies
0
Likes
6
Followers
44
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Piccolo express
Deborah Martuch, MT(ASCP), CPP
over 2 years ago
30262
Deborah Martuch, MT(ASCP), CPP
over 2 years ago
Quantra
Ashlee Byquist
over 2 years ago
50339
Ashlee Byquist
over 2 years ago
PPM Procedures/Resources
Jennifer Toncray
over 2 years ago
50256
Jennifer Toncray
over 2 years ago